by
David Wadlow
Basking Ridge, NJ 07920 USA
This article describes the applications, performance characteristics, mode and theory of operation, calibration, installation and maintenance procedures and the design and construction of axial turbine flowmeters, including two dual rotor axial turbine designs, propeller meters and spirometers. The article also includes descriptions of insertion axial turbines and multi-jet turbines.
[h=Axial turbine flowmeters]1[/h] The modern axial turbine flowmeter, when properly installed and calibrated, is a reliable device capable of providing the highest accuracies attainable by any currently available flow sensor for both liquid and gas volumetric flow measurement. It is the product of decades of intensive innovation and refinements to the original axial vaned flowmeter principle first credited to Woltman in 1790, and at that time applied to measuring water flow. The initial impetus for the modern development activity was largely the increasing needs of the US natural gas industry in the late 1940's and 50's for a means to accurately measure the flow in large diameter, high pressure, interstate natural gas lines. Today, due to the tremendous success of this principle, axial turbine flowmeters of different and often proprietary designs are used for a variety of applications where accuracy, reliability and rangeability are required in numerous major industries besides water and natural gas including oil, petrochemical, chemical process, cryogenics, milk and beverage, aerospace, biomedical and others.
Figure 1. Longitudinal section of an axial turbine flowmeter depicting the key components. The flowmeter body is usually a magnetically transparent stainless steel such as 304. Common end-fittings include face flanges (depicted), various threaded fittings and tri-clover fittings. The upstream and downstream diffusers are the same in bi-directional meters, and generally supported by 3 or more flat plates, or sometimes tubular structures, aligned with the body and which also act as flow straighteners. The relative size of the annular flow passage at the rotor varies among different designs. Journal rotor bearings are frequently used for liquids, while ball bearings are often used for gases. Magnetic reluctance pickups (depicted) are frequently used. Others types include mechanical and modulated carrier pickups. (1) End fitting - flange type shown; (2) Flowmeter body; (3) Rotation pickup - magnetic reluctance type shown; (4) Permanent magnet; (5) Pickup coil wound on pole piece; (6) Rotor blade; (7) Rotor hub; (8) Rotor shaft bearing - journal type shown; (9) Rotor shaft; (10) Diffuser support and flow straightener; (11) Diffuser; (12) Flow conditioning plate (dotted) - optional with some meters.
Figure 1 is a schematic longitudinal section through the axis of symmetry depicting the key components of a typical meter. As one can see, the meter is an inline sensor comprising a single turbine rotor, concentrically mounted on a shaft within a cylindrical housing through which the flow passes. The shaft or shaft bearings are located by end supports inside suspended upstream and downstream aerodynamic structures called diffusers, stators or simply cones. The flow thus passes through an annular region occupied by the rotor blades. The blades, which are usually flat but may be slightly twisted, are inclined at an angle to the incident flow velocity and hence experience a torque which drives the rotor. The rate of rotation, which can be up to several X 104 RPM for smaller meters, is detected by a pickup, which is usually a magnetic type, and registration of each rotor blade passing infers the passage of a fixed volume of fluid.
[h=General performance characteristics]2[/h] Axial turbines perform best when measuring clean, conditioned, steady flows of gases and liquids with low kinematic viscosities (below about 10-5 m2s-1, 10 cSt, although they are used up to 10-4 m2s-1, 100 cSt), and are linear for subsonic, turbulent flows. Under these conditions the inherent mechanical stability of the meter design gives rise to excellent repeatability performance. Not including the special case of water meters, which are described later, the main performance characteristics are:
  • Sizes, (internal diameter) range from 6 to 760 mm , (1/4'' - 30'').
  • Maximum measurement capacities range from 0.025 Am3/hr to 25,500 Am3/hr, (0.015 ACFM to 15,000 ACFM), for gases and 0.036 m3/hr to 13,000 m3/hr, (0.16 gpm to 57,000 gpm or 82,000 barrels per hour), for liquids, where A denotes actual.
  • Typical measurement repeatability is 0.1% of reading for liquids and 0.25% for gases with up to 0.02% for high accuracy meters. Typical linearities, (before electronic linearization), are between 0.25% to 0.5% of reading for liquids and 0.5% and 1.0% for gases. High accuracy meters have linearities of 0.15% for liquids and 0.25% for gases, usually specified over a 10:1 dynamic range below maximum rated flow. Traceability to NIST (National Institute of Standards and Technology) is frequently available allowing one to estimate the overall absolute accuracy performance of a flowmeter under specified conditions. Under ideal conditions, absolute accuracies for optimum designs and installations can come close to the accuracy capabilities at the NIST, which are stated as 0.13% for liquid flows and 0.25% for air.
  • Rangeability, when defined as the ratio of flow rates over which the linearity specification applies, is typically between 10:1 and 100:1.
  • Operating temperature ranges span -270oC to 650oC, (-450oF to 1200oF).
  • Operating pressure ranges span coarse vacuum to 414 MPa, (60,000 psi).
  • Pressure drop at the maximum rated flow rate ranges from around 0.3 kPa (0.05 psi) for gases to in the region of 70 kPa (10 psi) for liquids.
[h=Theory]2[/h] There are two approaches described in the current literature for analyzing axial turbine performance. The first approach describes the fluid driving torque in terms of momentum exchange, while the second describes it in terms of aerodynamic lift via airfoil theory. The former approach has the advantage that it readily produces analytical results describing basic operation, some of which have not appeared via airfoil analysis. The latter approach has the advantage that it allows more complete descriptions using fewer approximations. However, it is mathematically intensive and leads rapidly into computer generated solutions. One prominent pioneer of the momentum approach is Lee [1] who, using this approach, later went on to invent one of the few, currently successful, dual rotor turbine flowmeters, while Thompson and Grey [2] provided one of the most comprehensive models currently available using the airfoil approach, which for instance, took into account blade interference effects. In the following, I have used the momentum exchange approach to highlight the basic concepts of the axial turbine flowmeter.
In a hypothetical situation, where there are no forces acting to slow down the rotor, it will rotate at a speed which exactly maintains the fluid flow velocity vector at the blade surfaces.
Figure 2. Vector diagram for a flat-bladed axial turbine rotor. The difference between the ideal (subscript i) and actual tangential velocity vectors is the rotor slip velocity and is caused by the net effect of the rotor retarding torques. This gives rise to linearity errors and creates swirl in the exit flow. V incident fluid velocity vector; VE exit fluid velocity vector; exit flow swirl angle due to rotor retarding torques; blade pitch angle, same as angle of attack for parallel flow; rotor angular velocity vector; r rotor radius vector; F flow induced drag force acting on each blade surface; c blade chord; s blade spacing along the hub; c/s rotor solidity factor.
Figure 2 is a vector diagram for a flat bladed rotor with a blade pitch angle equal to . Assuming that the rotor blades are flat and that the velocity is everywhere uniform and parallel to the rotor axis, then referring to figure 2:
When one introduces the total flow rate this becomes:
where is the 'ideal' rotational speed, Q is the volumetric flow rate, A is the area of the annular flow cross section and is now the root-mean-square of the inner and outer blade radii, (R, a). Eliminating the time dimension from the left hand side quantity reduces it to the number of rotor rotations per unit fluid volume, which is essentially the flowmeter K factor specified by most manufacturers. Hence, according to Eq. (2), in the ideal situation the meter response is perfectly linear and determined only by geometry. (In some flowmeter designs the rotor blades are helically twisted to improve efficiency. This is especially true of blades with large radius ratios, (R/a). If the flow velocity profile is assumed to be flat, then the blade angle in this case may be described by tan =constant X r. This is sometimes called the 'ideal' helical blade.) In practice, there are instead a number of rotor retarding torques of varying relative magnitudes. Under steady flow the rotor assumes a speed which satisfies the following equilibrium:
Fluid driving torque = rotor blade surfaces fluid drag torque + rotor hub and tip clearance fluid drag torque + rotation sensor drag torque + bearing friction retarding torque (3)
Referring again to figure 2, the difference between the actual rotor speed, r, and the ideal rotor speed, , is the rotor slip velocity due to the combined effect of all the rotor retarding torques as described in Eq. (3), and as a result of which the fluid velocity vector is deflected through an exit or swirl angle, . Denoting the radius variable by r, and equating the total rate of change of angular momentum of the fluid passing through the rotor to the retarding torque, one obtains:
which yields:
where is the fluid density and NTis the total retarding torque. Combining Eqs. (1) and (4) and rearranging, yields:
The trends evident in Eq. (5) reflect the characteristic decline in meter response at very low flows and why lower friction bearings and lower drag pickups tend to be used in gas versus liquid applications and small diameter meters. In most flowmeter designs, especially for liquids, the latter three of the four retarding torques described in Eq. (3) are small under normal operating conditions compared with the torque due to induced drag across the blade surfaces. As shown in figure 2, the force, F, due to this effect acts in a direction along the blade surface and has a magnitude given by:
where CD is the drag coefficient and S is the blade surface area per side. Using the expression for drag coefficient corresponding to turbulent flow, selected by Pate et al. [3] and others, this force may be estimated by:
where Re is the flow Reynolds number based on the blade chord shown as dimension c in figure 2. Assuming is small compared with , then after integration, the magnitude of the retarding torque due to the induced drag along the blade surfaces of a rotor with n blades is found to be:
Combining Eqs. (7) and (5), and rearranging yields:
Eq. (8) is an approximate expression for K factor because it neglects the effects of several of the rotor retarding torques, and a number of important detailed meter design and aerodynamic factors, such as rotor solidity and flow velocity profile. Nevertheless, it reveals that linearity variations under normal, specified operating conditions are a function of certain basic geometric factors and Reynolds number. These results reflect general trends which influence design and calibration. Additionally, the marked departure from an approximate (actually via Re in Eq.(6)) dependence of the fluid drag retarding torque on flow properties under turbulent flow, to other relationships under transitional and laminar flow, gives rise to major variations in the K factor versus flow rate and media properties for low flow Reynolds numbers. This is the key reason why axial turbine flowmeters are generally recommended for turbulent flow measurement.
[h=Calibration, installation and maintenance]2[/h] Axial turbine flowmeters have a working dynamic range of at least 10:1 over which the linearity is specified. The maximum flow rate is determined by design factors related to size versus maximum pressure drop and maximum rotor speed. The minimum of the range is determined by the linearity specification itself. Due to small, unavoidable, manufacturing variances, linearity error curves are unique to individual meters and are normally provided by the manufacturer. However, although recommended where possible, the conditions of the application cannot usually and need not necessarily duplicate those of the initial or even subsequent calibrations. This has pivotal importance in applications where actual operating conditions are extreme or the medium is expensive or difficult to handle.
Figure 3. A typical single rotor axial turbine linearity error, or calibration, curve for a low viscosity fluid showing the main alternative presentations in current use. Higher accuracy specifications usually correspond to a 10:1 flow range down from Qmax, while extended operating ranges usually correspond to reduced accuracies. The hump in the depicted curve is a characteristic feature caused by flow velocity profile changes as Re approaches the laminar region. This feature varies in magnitude between meters. Sensitivity and repeatability performance degrades at low Re. Percent registration is only used with meters which have mechanical pickups. All other meters have a K factor. UVC and Re calibrations remain in effect at different known media viscosities provided Re or f/v stays within the specified range. Re is referenced to the connecting conduit diameter and is less within the flowmeter. The Re range shown is therefore approximate and can vary by an order of magnitude depending on the meter. Linearity error may also be expressed in terms of Strouhal number (fD/V) versus Re (VD/v) or Roshko number (fD2/v ), when instead D is a flowmeter reference diameter, [4].
Figure 3 depicts a typically shaped calibration curve of linearity versus flow rate expressed in terms of multiple alternative measures, various combinations of which may be found in current use. The vertical axis thus represents either the linearity error as a percentage of flow rate, a K factor expressed in terms of the number of pulses from the rotation sensor output per volume of fluid or the deviation from 100% registration; the latter only applying to flowmeters with mechanical pickups. The horizontal axis may be expressed in terms of flow rate in volume units/time, Reynolds number, (Re), or pulse frequency (from the rotation sensor for non-mechanical) divided by kinematic viscosity, (f/v), in units of Hz per m2s-1 , (Hz/cSt or Hz/SSU; 10-6 m2s-1 = 1 centistoke 31.0 seconds Saybolt Universal), and where kinematic viscosity is the ratio of absolute viscosity (u) to density. Calibrations are preferably expressed versus Re or f/v, which is proportional to Re. The hump shown in the curve is a characteristic frequently observed at lower Re and is due to velocity profile effects. K factor versus f/v calibration curves are specifically called universal viscosity curves (UVC) and for most meters are available from the manufacturer for an extra charge. A key utility of UVC is that where media type and properties differ significantly from those of the original calibration, accuracies much greater than the overall linearity error can still readily be obtained via the flowmeters UVC if the kinematic viscosity of the application is known. An alternative, advanced calibration technique [4], is to provide response in terms of Strouhal number versus Re or Roshko number. This approach is not widely adopted, but it is particularly relevant to high accuracy and extreme temperature applications because it further allows correct compensation for flowmeter thermal expansion errors.
The accuracy of axial turbine flowmeters is reduced by unconditioned flow, especially swirl. An installation incorporating flow conditioners along with specific upstream and downstream straight pipe lengths is generally recommended, [5]. Some axial turbine flowmeters can be purchased with additional large flow straighteners that mount directly ahead of the flowmeter body or conditioning plates which are integral to the body. The manufacturer is the first source of information regarding installation. Errors due to flow velocity pulsations are another concern, particularly in certain gas installations. However no standard technique for effectively counteracting this source of error has yet been adopted. Periodic maintenance, testing and recalibration is required because the calibration will shift over time due to wear, damage or contamination. For certain applications, especially those involving custody transfer of oil and natural gas, national standards, international standards and other recommendations exist which specify the minimum requirements for turbine meters with respect to these aspects, [6, 7, 8, 9, 10].
[h=Design and construction]2[/h] There are numerous, often proprietary, designs incorporating variations in rotors, bearings, pickups and other components in format and materials which are tailored to different applications. Meter bodies are available with a wide range of standard end-fittings. Within application constraints, the primary objective is usually to optimize the overall mechanical stability and fit in order to achieve good repeatability performance. Design for performance, application and manufacture considerations impact every internal component, but most of all the rotor with respect to blade shape and pitch, blade count, balance and rigidity versus, drag, stress and inertia, bearings with respect to precision versus friction, speed rating and durability and rotation pickup versus performance and drag.
Most low radius ratio blades are machined flat, while high ratio blades tend to be twisted. The blade count varies from about 6 to 20 or more depending on the pitch angle and blade radius ratio so that the required rotor solidity is achieved. Rotor solidity is a measure of the 'openness' to the flow such that higher solidity rotors are more highly coupled to the flow and achieve a better dynamic range. The pitch angle, which primarily determines the rotor speed, is typically 30o to 45o but may be lower in flowmeters designed for low density gas applications. Rotor assemblies are usually a close fit to the inside of the housing. In large diameter meters the rotor often incorporates a shroud around the outer perimeter for enhanced stability. Also, since large meters are often used for heavy petroleum products, via selection of a suitable wall clearance, the fluid drag resulting from this clearance gap is often designed to offset the tendency at high media viscosities for the meter to speed up at lower Reynolds numbers. The materials of construction range from non magnetic to magnetic steels to plastics.
Stainless steel ball bearings tend to be used for gas meters and low lubricity liquids such as cryogenic liquids and freon, while combination tungsten carbide or ceramic journal and thrust bearings are often considered best for many other liquid meters depending on the medium lubricity. Fluid bearings (sometimes called 'bearingless' designs) are often used in conjunction with the latter, but also sometimes with gases, for reducing the drag. They operate by various designs which use flow induced forces to balance the rotor away from the shaft ends. Bearing lubrication is either derived from the metered medium or an internal or external system is provided. The more fragile, jeweled pivot bearings are also used in certain gas applications and small meters. Sanitary meters may incorporate flush holes in the bearing assembly to meet 3A ----- and crevice standards.
The most common types of rotation sensor are magnetic, modulated carrier and mechanical, while optical, capacitative and electrical resistance are also used. In research, a modulated nuclear radiation flux rotation sensor for use in certain nuclear reactors has also been reported, [11, 12]. Mechanical pickups, which sometimes incorporate a magnetic coupling, are traditional in some applications, can have high resolution and one advantage that they require no electrical power. However the pickup drag tends to be high. The magnetic and modulated carrier types utilize at least a coil in a pickup assembly which screws into the meter housing near the rotor. In magnetic inductance types, which are now less common, the blades or shroud carry magnetized inserts, and signals are induced in the coil by the traversing magnetic fields. In the more prevalent magnetic reluctance type, an example of which is schematically depicted in figure 1, the coil is wrapped around a permanent magnet or magnet pole piece in the pickup assembly which is mounted next to a high magnetic permeability bladed rotor (or machined shroud). The latter is then typically made of a magnetic grade of stainless steel such as 416, 430 or 17-4Ph. As the rotor turns, the reluctance of the magnetic circuit varies producing signals at the coil. In the more expensive modulated carrier types, the rotor need only be electrically conductive. The coil is part of a radio frequency, (RF), oscillator circuit and proximity of the rotor blades changes the circuit impedance giving rise to modulation at a lower frequency which is recovered. The RF types have much lower drag, higher signal levels at low flow and can operate at temperatures above the Curie point of typical ferromagnetic materials. They are preferred for wide dynamic range and high temperature applications. Bi-directional flowmeters, usually have two magnetic pickups to determine flow direction. This is useful, for example, in the monitoring of container filling and emptying operations often encountered in sanitary applications. Multiple magnetic pickups are also used in some designs to provide increased measurement resolution. Regarding output, various pulse amplifiers, totalizers, flow computers for gas pressure and temperature correction, along with 4-20 mA and other standard interface protocols, are available to suit particular applications. As an example of advanced transmitters, at least one manufacturer provides a real-time, miniature, reprogrammable, 'smart' transmitter which is integrated into the pickup housing along with a meter body temperature sensor, for full viscosity compensation and UVC linearization. These are for use in dedicated applications, such as airborne fuel management, where the medium viscosity-temperature relationship is known.
Certain applications have uniquely different design requirements and solutions, and two are discussed separately in the following.
[h=Propeller meters]3[/h] Propeller meters are used in either municipal, irrigation or waste water measurement. Although in some designs propeller and turbine meters look almost identical and operate on the same axial rotor principle, this type of flowmeter is currently commercially and officially, [13, 14], distinguished as a separate category distinct from the axial turbine. Diameters of up to 2440 mm (96'') are available. The flow rate capacity of a 1800 mm (72") diameter propeller meter is up to about 25,000 m3/hr, (110,000 gpm). Typical accuracies are 2% of reading. A primary requirement is ruggedness, and it is in the designs most suited to harsh environments that the formats are most distinctive. Rotor and pickup assemblies are generally flanged to the housing and removable. The rotors have large clearances, are often cantilevered into the flow, and supported via a sealed bearing without stators. The rotors are typically made of plastic or rubber and carry as few as 3 highly twisted, high radius ratio blades. Pickups are always mechanical and frequently have magnetic couplings.
[h=Spirometers]3[/h] Monitoring spirometers measure the volumes of gas flows entering and leaving the lungs and may also be incorporated in ventilator circuits. Diagnostic spirometers are used to monitor the degree and nature of respiration. With these a clinician may determine patient respiratory condition by various measures and clinical maneuvers. Low cost, light weight, speed of response and patient safety are major considerations. Measurement capabilities include the gas volume of a single exhalation and also the peak expiratory flow for diagnostic types, measured in liters and liters per second, respectively. Various technologies are used. However, the Wright respirometer, named after the original inventor [15], today refers to a type of hand-held monitoring spirometer which utilizes a special type of tangential turbine transducer with a two bladed rotor connected to a mechanical pickup and a dial readout for the volume. These particular spirometers are routinely used by respiratory therapists for patient weaning and ventilator checking. Other axial turbine based flowmeters are available for ventilation measurements involving, for example, patient metabolics measurements. One axial turbine based diagnostic spirometer made by Micro Medical, Ltd utilizes an infrared, optical pickup and has a battery powered microprocessor controlled display. In these medical devices, rotors tend to be plastic with a large blade radius ratio. Flow conditioning is minimal or absent. The meters are typically accurate to a few percent of reading. In the U.S. spirometers are designated as class 2 medical devices and as such certain FDA approvals are required concerning manufacture and marketing. In the EU they are class IIb medical devices under a different system and other approvals are required.
[h=Dual rotor axial turbines]2[/h] Dual rotor axial turbines have performance features not found in single rotor designs.
In 1981 Lee et al. [16] were issued a US patent for a self-correcting, self-checking dual rotor turbine flowmeter which is currently manufactured exclusively by Equimeter, Inc. and sold as the Auto-Adjust. This is a high accuracy flowmeter primarily intended for use on large natural gas lines where even small undetected flow measurement errors can be costly. It incorporates two closely coupled turbine rotors which rotate in the same direction. The upstream rotor is the main rotor and the second rotor, which has a much shallower blade angle, is the sensor rotor. Continuous and automatic correction of measurement errors due to varying bearing friction is achieved by calculating the flow rate based on the difference between the rotor speeds. As shown in figure 2 and discussed in the theory section, the flow exit angle is due to the net rotor retarding torque. If this torque increases in the main rotor, thereby reducing its speed, the exit angle increases and the speed of the sensor rotor is then also reduced. The meter is also insensitive to inlet swirl angle, because the swirl affects both rotor speeds in the same sense and the effect is then subtracted in the flow calculation. The meter also checks itself for wear and faults by monitoring the ratio of the two rotor speeds and comparing this number with the installation value, [17].
A dual rotor liquid flowmeter, invented by Ruffner et al. [18], was introduced by Exact Flow, LLC. It is offered as a high accuracy flowmeter, (up to 0.1% linearity and 0.02% repeatability), which has an extraordinarily wide dynamic range of 500:1 with a single viscosity liquid. This flowmeter had early commercial success in fuel flow measurement in large jet engine test stands where the wide dynamic range is particularly useful, [19]. The meter comprises two, closely and hydraulically coupled rotors which rotate in opposite directions. Due to the exit angle generated by the first rotor, the second rotor continues to rotate to much lower flow rates compared with the first.
[h=Two-phase flow measurement using axial turbines]2[/h] A differential pressure producing flowmeter such as a venturimeter in series with a turbine is known to be a technically appropriate and straightforward method for measuring the volumetric and mass flow rates of some fine, solid aerosols. However, this section highlights a current research area in the application of axial rotor turbine meters to a range of industrial flow measurement problems where gas/liquid, two phase flows are encountered. Customarily turbine meters are not designed for and cannot measure such flows accurately. Errors of the order 10% arise in metering liquids with void fractions of around 20%. Such flows are normally measured after gas separators. Although this problem is not restricted to these industries, the current main impetuses for the research are the direct measurement of crude oil in offshore multiphase production systems, the measurement of water/steam mixtures in the cooling loops of nuclear reactors and the measurement of freon liquid-vapor flows in refrigeration and air conditioning equipment. Several techniques investigated so far use an auxiliary sensor. This may either be a void fraction sensor or a pressure drop device such as a venturimeter or drag disk, of which the pressure drop approach appears to be technically more promising, [20, 21]. Also from a practical standpoint, gamma densitometers for measuring void fraction are additional and expensive equipment and not for instance, well adapted for use in undersea oil fields. Two techniques currently studied do not require an auxiliary in-line sensor. The first uses the turbine meter itself as the drag body and combines the output of the turbine with that of a small differential pressure sensor connected across the inlet and outlet regions. This technique requires a homogenizer ahead of the turbine and measurement accuracies of 3% for the volumetric flow rates of both phases have recently been reported for air/water mixtures up to a void fraction of 80%, [22]. The second technique is based entirely on analysis of the turbine output signal and has provided significant correlations of the signal fluctuations with void fraction. Accuracies of water volumetric flow rate measurement of 2% have been reported when using air/water mixtures with void fractions of up to 25%, [23].
[h=Insertion axial turbine flowmeters]1[/h] These flowmeters comprise a small axial rotor mounted on a stem which is inserted radially through the conduit wall, often through a shut-off valve. They measure the flow velocity at the rotor position from which the volumetric flow rate is inferred. They are an economical solution to flow measurement problems where pipe diameters are high and accuracy requirements are moderate, and also may be technically preferred where negligible pressure drop is an advantage, as in high speed flows. They are typically more linear than insertion tangential turbine flowmeters and compete also with magnetic and vortex shedding insertion flowmeters. They are available for the measurement of a range of liquids and gases, including steam, similar to the media range of full bore axial turbines, and have a similarly linear response. The rotors, which are usually metal but can be plastic, typically have diameters of 25 mm to 51 mm (1'' to 2''). They can be inserted into pipes with diameters ranging from 51 mm to 2032 mm (2'' to 80''). Velocity measurement ranges cover 0.046 m.s-1 to 91 m.s-1, (9 fpm to 18,000 fpm) for gases and 0.03 m.s-1 to 30 m.s-1 (6 fpm to 6000 fpm) for liquids. Dynamic ranges vary between 10:1 and 100:1. The maximum flow rate measurement capacity in a 1836 mm (72'') diameter pipe can be as high as nearly 56,500 m3/hr, (about 250,000 gpm). Since these devices are local velocity sensors, calculating the volumetric flow rate requires a knowledge of the area velocity profile and the actual flow area. Flow conditioning is therefore particularly important for accurate volumetric measurements, while radial positioning, which is a further responsibility of the user, must be according to the manufacturer's recommendation, which can either be centerline, one third of the diameter, 12% of the diameter or determined by 'profiling'. Quick [24] discusses operation and installation for natural gas measurement. Although linearities or 'accuracies' may be quoted up to 1% of velocity, achieving the same accuracy for the volumetric flow rate, although possible, may be difficult or impractical. In this respect a unique dual rotor design, exclusive to Onicon, Inc., and primarily used for chilled water flow measurement in HVAC systems, requires less flow conditioning than single rotor designs. It comprises two rotors which rotate in opposite directions. The output is based on the average rotor speed. Any flow swirl present due to poor flow conditioning changes the speed of rotation of each rotor by the same but opposite amounts. Swirl induced error is thus virtually absent in the averaged output. Also, flow profile sampling is improved over that of a single rotor. The devices are calibrated using a volumetric prover and the specified accuracy of 2% of reading is for volumetric flow rate rather than velocity. This is the total error and includes an allowance for dimensional variations in industry standard pipes.
[h=Multi-jet turbine flowmeters]1[/h] These are linear, volumetric flowmeters designed for liquids measurements and comprise a single, radial-vaned impeller, vertically mounted on a shaft bearing within a vertically divided flow chamber, sometimes called a distributor. The impeller is often plastic and may even be neutrally buoyant in water. There are various designs, however typically both chambers access a series of radially distributed and angled jets. The lower chamber jets connect to the flowmeter input port and distribute the flow tangentially onto the lower region of the impeller blades, while the upper series, which is angled oppositely, allow the flow to exit. The flow pattern within the flow chamber is thus a vertical spiral and the dynamic pressure drives the impeller to track the flow. This design gives the meters good sensitivity at low flow rates. Due to the distribution principle the meters are also insensitive to upstream flow condition. Impeller rotation pickups are always mechanical, often magnetically coupled, and frequently also connect with electric contact transmitters. They are primarily used in water measurement including potable water measurement for domestic and business billing purposes and in conjunction with energy management systems such as hot water building or district heating, and to a much lesser extent in some chemical and pharmaceutical industries for dosing and filling systems involving solvents, refrigerants, acids and alkalis with absolute viscosities less than 4.5 mPa.s, (0.045 Poise). Available sizes range from 15 mm to 50 mm. Dynamic ranges lie between 25:1 and 130:1 and flow measurement ranges cover 0.03 to 30 m3/hr, (0.13 to 130 gpm). Measurement linearities range between 1% to 2%, with typical repeatabilities of 0.3%. Operating temperatures range from normal to 90oC (200oF) and maximum operating pressures are available up to 6.9 MPa, (1000 psi). A number of potable water measurement systems come with sophisticated telemetry options which allow remote interrogation by radio or telephone. For potable water applications in the U.S. these meters normally comply with the applicable AWWA standard, [25], while in Europe EEC, DIN and other national standards apply.
In the author's opinion, there is also another type of vaned flowmeter which could be classified as a type of multi-jet turbine. This type comprises an axially mounted, vaned impeller with an upstream element which imparts a helical swirl to the flow. The transducer is typically a small, low cost, sometimes disposable, plastic component, and is usually designed for liquids, (but also to lower accuracies, gases), low flow rate measurements, (down to 50ml/min). The dynamic range is high and accuracies range up to 0.5%. Goss [26] describes one particular design in current use. Specialized applications cover the pharmaceutical, medical and beverage industries.
[h=Conclusion]1[/h] However anachronistic intricate mechanical sensors may appear amidst current everyday high technology, there are fundamental reasons why axial turbines are likely to experience continued support and development rather than obsolescence, especially for in-line applications requiring in the region of tenth percent volumetric accuracy. Mechanical coupling is the most direct volume interaction for a flowing fluid, which is why mechanical meters historically developed first and continue to be the most accurate and reliable types of flowmeters for so many different fluids. Other, non-mechanical, perhaps higher technology, or newer approaches thus face high demands for accurate compensation to render such less directly volume-coupled techniques as generally accurate, or more accurate. This is because the error in each corrected factor or assumption in an indirect technique contributes to the overall error. The technology of high accuracy flowmeters continues to be driven by applications, such as the custody transfer of valuable oil and natural gas, which demand high accuracy and reliability. There is a continuing demand for accurate and reliable water flowmeters. By reason of long proven field experience, turbine and other vane type devices have become one of a few broadly accepted techniques in many major applications such as these where the demands for flow sensors is significant or growing.
[h=References]1[/h] 1. W. F. Z. Lee, and H. J. Evans, Density effect and Reynolds number effect on gas turbine flowmeters, Trans. ASME, J. Basic Eng., 87 (4): 1043-1057, 1965.
2. R. E. Thompson, and J. Grey, Turbine flowmeter performance model, Trans. ASME, J. Basic Eng., 92 (4): 712-723, 1970.
3. M. B. Pate, A. Myklebust, and J. H. Cole, A computer simulation of the turbine flow meter rotor as a drag body, Proc. Int. Comput. in Eng. Conf. and Exhibit 1984, Las Vegas: 184-191, NY, NY: ASME, 1984.
4. P. D. Olivier, and D. Ruffner, Improved turbine meter accuracy by utilization of dimensionless data, Proc. 1992 Nat. Conf. Standards Labs. (NCSL) Workshop and Symp.: 595-607, 1992.
5. ISA-RP 31.1, Specification, installation and calibration of turbine flowmeters, Research Triangle Park, NC: ISA, 1977.
6. ANSI/ASME MFC-4M-1986 (R1990), Measurement of gas flow by turbine meters, NY, NY:ASME.
7. API MPM, Ch. 5.3, Measurement of liquid hydrocarbons by turbine meters, 3rd Ed., Washington, DC:API (Amer. Petroleum Inst.), 1995.
8. AGA Transmission Meas. Committee Rep. No. 7, Measurement of fuel gas by turbine meters, Arlington, VA:AGA (Amer. Gas Assoc.), 1981.
9. Int. Recommendation R32, Rotary piston gas meters and turbine gas meters, Paris, Fr.: OIML (Int. Organization of Legal Metrology), 1989.
10. ISO 9951:1993, Measurement of gas flow in closed conduits - turbine meters, Geneva, Switz.: Int. Organization for Standardization, (also available ANSI), 1993.
11. T. H. J. J. Van Der Hagen, Proof of principle of a nuclear turbine flowmeter, Nucl. Technol., 102 (2): 167-176, 1993.
12. K. Termaat, W. J. Oosterkamp, and W. Nissen, Nuclear turbine coolant flow meter, U.S. Patent 5,425,064, 1995.
13. AWWA C704-92, Propeller-type meters for waterworks applications, Denver, CO: Amer. Water Works Assoc., 1992.
14. ANSI/AWWA C701-88, Cold water meters - turbine type, for customer service, Denver, CO: Amer. Water Works Assoc., 1988.
15. B. M. Wright, and C. B. McKerrow, Maximum forced expiratory flow rate as a measure of ventilatory capacity, Brit. Med. J.: 1041-1047, 1959.
16. W. F. Z. Lee, R. V. White, F. M. Sciulli, and A. Charwat, Self-correcting self-checking turbine meter, U.S. Patent 4,305,281, 1981.
17. W. F. Z. Lee, D. C. Blakeslee, and R. V. White, A self-correcting and self-checking gas turbine meter, Trans. ASME, J. Fluids Eng., 104: 143-149, 1982.
18. D. F. Ruffner, and P. D. Olivier, Wide range, high accuracy flow meter, U.S. Patent 5,689,071, 1997.
19. D. F. Ruffner, Private communication, 1996.
20. A. Abdul-Razzak, M. Shoukri, and J. S. Chang, Measurement of two-phase refrigerant liquid-vapor mass flow rate - part III: combined turbine and venturi meters and comparison with other methods, ASHRAE Trans.: Research 101 (2): 532-538, 1995.
21. W. J. Shim, T. J. Dougherty, and H. Y. Cheh, Turbine meter response in two-phase flows, Proc. Int. Conf. Nucl. Eng. - 4, 1 part B: 943-953, NY, NY:ASME, 1996.
22. K. Minemura, K. Egashira, K. Ihara, H. Furuta, and K. Yamamoto, Simultaneous measuring method for both volumetric flow rates of air-water mixture using a turbine flowmeter, Trans. ASME, J. Energy Resources Technol., 118: 29-35, 1996.
23. M. W. Johnson, and S. Farroll, Development of a turbine meter for two-phase flow measurement in vertical pipes, Flow Meas. Instrum., 6 (4): 279-282, 1995.
24. L. A. Quick, Gas measurement by insertion turbine meter, Proc. 70th Int. School Hydrocarbon Meas., OK, 1995. (Available E. Blanchard, Arrangements Chair, Shreveport, LA, 318 868 0603.)
25. ANSI/AWWA C708-91, Cold-water meters, multi-jet-type, Denver, CO: Amer. Water Works Assoc., 1991.
26. J. Goss, Flow meter, U.S. Patent 5,337,615, 1994.
27. C. R. Sparks, Private communication, 1996.
[h=Further information]1[/h] D. W. Spitzer (ed.), Flow measurement, Research Triangle Park, NC: ISA, 1991, is a popular 646 page practical engineering guide of which four chapters concern turbine flowmeters, sanitary flow meters, insertion flow meters and custody transfer issues.
A. J. Nicholl, Factors affecting the performance of turbine meters, Brown Bov. Rev., 64 (11): 684-692, 1977, describes some basic design factors not commonly discussed elsewhere.
J. W. DeFeo, Turbine flowmeters for measuring cryogenic liquids, Adv. Instrum. Proc., 47 pt.1: 465-472, ISA, 1992, provides guidance for a less frequently discussed application in which axial turbines perform well.
ATS Standardization of spirometry, 1994 Update, Am. J. Respir. Care Med. 152: 1107-1136, 1995, is the latest version of the official US guideline for spirometry generated by the American Thoracic Society.
M. D. Lebowitz, The use of expiratory flow rate measurements in respiratory disease, Ped. Pulmonology, 11: 166-174, 1991, provides a review of the diagnostic usefulness of PEFR using portable spirometers.
W. M. Jungowski, and M. H. Weiss, Effects of flow pulsation on a single-rotor turbine meter, Trans. ASME, J. Fluids Eng., 118 (1): 198-201, 1996.
C. R. Sparks, and R. J. McKee, Method and apparatus for assessing and quantifying pulsation induced error in gas turbine flow meters, U.S. Patent 5,481,924, 1996, assigned to the Gas Research Institute, Chicago, is a potential solution to the accurate measurement of pulsating gas flows which will require engineering development and a high performance rotation sensor [27].
K. Ogawa, S. Ito, C. Kuroda, Laminar-turbulent velocity profile transition for flows in concentric annuli, parallel plates and pipes, J. Chem. Eng. Japan, 13 (3): 183-188, 1990, provides mathematical descriptions of velocity profiles.
J. Lui, B. Huan, Turbine meter for the measurement of bulk solids flowrate, Powder Technol. 82: 145-151, 1995, describes theory and experiments relating to a very simple design for a unique application, namely the volumetric measurement of plug flows of sands in a pipe.