  1. Flowmeter for custody transfer

    Hello
    What are the required repeatability and linearity for a flowmeter used for custody transfer?
    Also what is "Accuracy" shown on shown on suppliers specification, as i read accuracy is only a qualitative term?

  2. # ADS
    Spons Circuit
    Join Date
    Always
    Posts
    Many
     
  3. Quote Originally Posted by hassanfadil View Post
    Hello
    What are the required repeatability and linearity for a flowmeter used for custody transfer?
    Also what is "Accuracy" shown on shown on suppliers specification, as i read accuracy is only a qualitative term?
    Salaam

    My understanding for the above is as follows:
    The meter minimum performance (accuracy & repeatability) criteria for a gas custody transfer application is defined in relevant insternational measurement standards.
    The inherent accuracy of a meter is often published in the manufacturer’s specification, and may be expressed as repeatability and/or linearity. In other words,
    accuracy is based on how repeatable and how linear the meter can stay within the manufacturer’s performance specifications.
    Manufacturers’ specifications are based on meter operation
    within recommended flow ranges, within a narrow
    range of pressures, temperatures, and fluid viscosities. For
    custody transfer applications, meters with the highest inherent
    accuracy should be used and should be proved (mostly for liquid applications)on site .
    The meters should operate within the manufacturer’s specifications.

  4.    Sponsored Links



    -

  5. Thanks for the informative reply
    Yes that's what am asking but my problems are:
    1) the API standard shows that Uncertainty should be +/- 0.027 with 95% level of confidence, which can be achieved by: The typical requirement is that a meter must repeat within +/- 0.05%
    in 5 consultative runs. For example, a repeatability of +/ - 0.12% in 10 consecutive runs or a repeatability of +/- 0.17% in 15 consecutive
    runs both meet +/- 0.027% at 95% confidence level.
    2) The API standard does not show the linearity requirement.

  6. Quote Originally Posted by hassanfadil View Post
    Thanks for the informative reply
    Yes that's what am asking but my problems are:
    1) the API standard shows that Uncertainty should be +/- 0.027 with 95% level of confidence, which can be achieved by: The typical requirement is that a meter must repeat within +/- 0.05%
    in 5 consultative runs. For example, a repeatability of +/ - 0.12% in 10 consecutive runs or a repeatability of +/- 0.17% in 15 consecutive
    runs both meet +/- 0.027% at 95% confidence level.
    2) The API standard does not show the linearity requirement.
    It's true that API do not define the linearity numbers that should go with the meter. (Probably the reason for this could be with the diffent physical properties for the liquid ( e.g.: phase, viscosity etc) & the variety of fluids fluids to meter (crude, lpg, lng,other refined products) it could be difficult to define a particular number for linearity). Hence for majority of liquid custody transfer applications, this linearity requirement is defined in the contract between buyer & seller with their mutual agreement. Secondly it also depends on the what r the best linearity meters available in the market. For a turbine meter / ultrasonic meter the linearity is typically +/-0.15%.

    I hope this answers your query.

  7. Thanks Shabbir2009 you've answered my question, and thanks again for the fast replies ^_^

  8.    Spons.


Similar Threads

  1. Coriolis flow for custody transfer
    By davidseins in forum Instrumentation & Control
    Replies: 2
    Last Post: 06-06-2013, 09:21 AM
  2. LNG Custody Transfer Handbook
    By s@ndy in forum Engineering Certificates
    Replies: 10
    Last Post: 10-14-2011, 05:52 PM
  3. Article: Flowmeter for custody transfer
    By Shabbir2009 in forum Home Articles
    Replies: 0
    Last Post: 03-24-2011, 05:57 AM
  4. Article on Natural Gas Custody Transfer Measurement
    By Shabbir2009 in forum Pipeline And Fluid Flow
    Replies: 0
    Last Post: 12-22-2010, 05:31 AM
  5. API MPMS Chap. 18 - Custody Transfer
    By sambun in forum Laboratory Measurements
    Replies: 9
    Last Post: 06-16-2010, 01:57 AM

Tags for this Thread

Bookmarks

Posting Permissions

  • You may not post new threads
  • You may not post replies
  • You may not post attachments
  • You may not edit your posts
  •  
1 2 3 4 5 6 7 8 9 10 11 12 13 14 15 16 17 18 19 20 21 22 23 24 25 26 27 28 29 30 31 32 33 34 35 36 37 38 39 40